Analogue Input Range Configuration.

U3 IO Analogue Input

  • Please log in to reply
2 replies to this topic

#1 RRachev

RRachev
  • Members
  • 4 posts

Posted 16 July 2014 - 02:59 AM

Hello,

 

I am trying to take a single ended reading of a voltage using a U3-LV LabJack, and programming it in C#. The line of code I use to do so is as follows:

 

                            LJUD.eAIN(u3.ljhandle, 4, 31, ref Rec_Volt, 0, 0, 0, 0);

 

However the voltage is in the range +-10, and the readings only go as far as 2.5v. I would like to adjust the range of the readings, and I am trying to do so with the following line:

 

                            LJUD.ePut(u3.ljhandle, LJUD.IO.PUT_AIN_RANGE, 4, LJUD.RANGES.BIP10V,0);

 

However it does not adjust giving me the following problem: "Argument 4: cannot convert from 'LabJack.LabJackUD.LJUD.RANGES' to 'double'." & "Argument 5: cannot convert from 'int' to 'double[]'."
 

Does anyone have an idea of how to properly configure the range, and get the desired readings.

 

Thanks in advance.



#2 LabJack Support

LabJack Support
  • Admin
  • 8677 posts

Posted 16 July 2014 - 02:57 PM

From LabJackUD.h the constant for BIP10V is 2, but ranges have no effect on a U3 anyway so it won't do anything.

 

The normal range on all 16 low-voltage analog inputs is 0-2.4V.  If you take a differential reading with negative channel 32 you get the special 0-3.6V range.

 

http://labjack.com/s...users-guide/2.6

 

http://labjack.com/s...s-guide/2.6.3.6

 

So you can use a resistive voltage divider or the LJTick-Divider-BIP10V.  Or switch to a U3-HV, U6, or T7.



#3 LabJack Support

LabJack Support
  • Admin
  • 8677 posts

Posted 16 July 2014 - 03:15 PM

Also, for a call like LJUD.IO.PUT_AIN_RANGE (which isn't used on the U3), the fourth parameter needs to be a double. Typecast BIP10V to a double to fix the "Argument 4: cannot convert .." error:

 

LJUD.ePut(u3.ljhandle, LJUD.IO.PUT_AIN_RANGE, 4, (double)LJUD.RANGES.BIP10V, 0);