Reset CNT, and counting impulses
#1
Posted 22 February 2013 - 11:44 AM
#2
Posted 22 February 2013 - 12:26 PM
#3
Posted 22 February 2013 - 12:27 PM
#4
Posted 23 February 2013 - 02:17 AM
#5
Posted 23 February 2013 - 02:28 AM

#6
Posted 23 February 2013 - 02:58 AM
#7
Posted 25 February 2013 - 09:50 AM
On our newer devices (U3/U6/UE9/T7), the digital inputs are pulled high (3.3V) internally. So to create counts you would tap a wire connected to GND. But on the U12, the digital inputs are pulled low internally, so to create counts you would tap a wire connected to +5V.What do you mean with "the digital I/O is normally low on the U12, so you would use a jumper from +5V to create test counts"?
If you want to send it back to us directly, we can take a look at it. Contact [email protected] to follow-up on that.What else can I try? Send it back to the store I bought it? But I donīt have warranty any more...
#8
Posted 20 August 2013 - 09:40 AM
The counter problem here was that the watchdog was stuck on. See the new item #6 on the U12 Hardware Troubleshooting page:
http://labjack.com/s...troubleshooting