i have a problem with the settings for Labjack u12 in Agilent vee.
I have to record water-pressure with a sensor. From this sensor voltage gets to the labjack and then to agilent vee.
Can you help me, which settings and where i have to type in, to get a Samplingrate of 500 per second?
I have the labjackdrivers from meilhaus electronics and chose the EAnalogIn. Is this the right one? Or should I choose StreamStart-StreamRead-StreamClear?
The next problem is, that i don't know how to write this data in a csv-file with a timestamp of milliseconds.
I hope you understand me. I'm from Germany ind my english isn't very good.
Thank you for your help!!!
fugger

Labjack in Agilent Vee 7.5
Started by
fugger
, Feb 14 2013 03:12 AM
1 reply to this topic
#1
Posted 14 February 2013 - 03:12 AM
#2
Posted 14 February 2013 - 01:17 PM
To acquire 1 channel at 500 scans/second, you will have to use stream mode:
http://labjack.com/u12/datarates
We are not experts at file I/O in VEE, so can't help much there, but you just need to use whatever basic file I/O is supported in VEE. As for the timestamp, you know the timestamp based on scan number. For example, if you are scanning at 500 scans/second, you know that the 1000th scan happened 2 seconds after the first scan.
http://labjack.com/u12/datarates
We are not experts at file I/O in VEE, so can't help much there, but you just need to use whatever basic file I/O is supported in VEE. As for the timestamp, you know the timestamp based on scan number. For example, if you are scanning at 500 scans/second, you know that the 1000th scan happened 2 seconds after the first scan.