EI1050 Labview example


  • Please log in to reply
2 replies to this topic

#1 Duser

Duser
  • Members
  • 2 posts

Posted 13 November 2012 - 05:07 AM

I have an EI1050 probe and it works fine with your program Labjack Python EI1050 example, but when I run labview example "U3 Single EI-1050 Example" I get this error "LabJack Error #52: CRC failed for SHT communication occured at LJUD_GetNextResult.vi" Why is this?

#2 LabJack Support

LabJack Support
  • Admin
  • 8677 posts

Posted 13 November 2012 - 08:10 AM

Strange. Try running with the lightbulb on, to slow down LabVIEW, and see if that makes a difference. I do see that the LabVIEW example uses FIO6 for Red/Brown, while the python utility uses FIO7, so make sure you move those wires. You could also try changing the LV example to use FIO7 the same as the python utility.

#3 Duser

Duser
  • Members
  • 2 posts

Posted 14 November 2012 - 01:44 AM

Thank you very much for your quick answer and sorry, I did not realize the change in FIO6 and FIO7. Now both of them work-