Command and response Vs Streaming


  • Please log in to reply
1 reply to this topic

#1 Hybrid Engineer

Hybrid Engineer
  • Members
  • 9 posts

Posted 25 October 2012 - 05:15 AM

Hi I am currently reading the analogue inputs using the command and response mode but would like to be able to read the channels faster. I have looked into streaming mode and was just wondering if at this speed the labjacks accuracy is maintained especially when reading thermocouples in uV. Thanks

#2 LabJack Support

LabJack Support
  • Admin
  • 8677 posts

Posted 25 October 2012 - 07:38 AM

Speed does not effect accuracy. The only thing to watch for is that increased speed can expose settling errors if your source has too high of source impedance. I would not expect a problem with a thermocouple, but you can do a quick test. Scan an extra channel right before your thermocouple, and jumper that channel to GND or VS to see if it makes a difference whether you are settling from 0V or 5V.